[3view] knife moving problem

Minh Huynh minh.huynh at sydney.edu.au
Fri Aug 24 01:21:43 CEST 2018


Hi everyone,

I have had some issues with knife cutting/moving since upgrading to GMS 3 although I’m not sure it’s the same problem as described below, we have a Zeiss Sigma 3View.

When I noticed the sample was no longer cutting I opened up the chamber and watched the knife, it appeared as though the cutting window had completely changed from the original position. I tried to replicate and record this using an eyepiece camera to send to Gatan however I couldn’t get it to happen again! At the time, I just assumed I made a mistake but since then it has happened a few more times. I just restarted everythjng - computer, digiscan sometimes SEM and things seem to be back to normal.

We’ve had some other more pressing issues since then relating to the new BSE OnPoint detector under performing  so I haven’t followed up on the issue above.

I hope that helps and look forward to seeing some of you at IMC19.

Regards,

Minh


Dr Minh Huynh | Multi-Modal Microscopist
Australian Centre for Microscopy & Microanalysis

THE UNIVERSITY OF SYDNEY
Rm 111A, Madsen Building F09 | The University of Sydney | NSW | 2006 | Australia
T +61 2 9351 7525 | F +61 2 9351 7682
E minh.huynh at sydney.edu.au | W http://sydney.edu.au/acmm
________________________________
From: 3view-bounces at open-em.org <3view-bounces at open-em.org> on behalf of Matt Russell <Matt.Russell at crick.ac.uk>
Sent: Friday, 24 August 2018 3:50:27 AM
To: Małgorzata Śliwińska
Cc: Mailing list for Gatan 3View Serial Block-Facing
Subject: Re: [3view] knife moving problem

Hi Gosia,

We’ve had the IO exception error you posted. In our case it was while operating the 3View remotely. To resolve this we just have to turn off the microscope PC completely and turn it back on (restart doesn’t reset the communications with the microscope, and it has to be turned off properly, it won’t work if you turn it on after a forced power off because the computer crashed). The microscope was fine, it’s just the computer not communicating with it properly.

We think the cause was a missing Default Gateway IP address in the IPv4 settings of the (3View PC, I think).

Best,

Matt

Matt Russell, PhD
Electron Microscopy STP
The Francis Crick Institute
1 Midland Road
London
NW1 1AT

T: +442037962006
E: matt.russell at crick.ac.uk<mailto:matt.russell at crick.ac.uk>
W: www.crick.ac.uk<https://protect-au.mimecast.com/s/DCN5COMxNytR3k0mHE2PEl?domain=crick.ac.uk>

On 23 Aug 2018, at 08:27, Małgorzata Śliwińska <m.sliwinska at nencki.gov.pl<mailto:m.sliwinska at nencki.gov.pl>> wrote:


Hi Ilya,

Thank you very much for answer, I refer to your comments in the text below:

Hi Gosia,

1.

I agree with Joergen, it is difficult to target the problem. Especially, since it happened only few times during the year. Your events also may be due to certain accidents, such as miscommunication between DM and the knife controller (perhaps, in the second case with the error message: 3View record aborted:Timeout moving knife position to NEAR).

I am on DM 2, you can be wright with the suggestion about miscommunication as I have some other problems (see attached pictures).



This  error  occurs more often (already more than 20 times!) and during the scan normally. Than the communicate at Gatan computer appears first (3View communicate picture) and immediately another on microscope (microscope communicate). When it happen I lose a control over the microscope, the console is dead, I can not change magnification or anything and it is impossible to turn off the EHT or go to HV/VP mode but the scan goes on as long as I do not stop it.  In this situation I have to turn off everything (with EHT on, and server gives error communicate-see attached), press the yellow button on microscope (standby mode) and after a while turn it on back on green button. This time the microscope computer do not starts automatically and I have to turn it on manually. Some popping up windows appear like that the computer can not read the EHT, and vacuum value and "Application SmartSem version 5.5 with service pack 6 08-April-13 failed to register successfully"



Normally I start the Smart Sem, turn off EHT, next turn on EHT again and open Digital micrograph. After the error like this DM shows no windows, which I normally have and I have to open them from the panel (floating windows layout). Sometimes there is no Digi Scan connected, so I restart the computer again.



2.

With the other symptoms, I also had a weird case when I suppose to cut but nothing was going on for several microns. Basically, I just opened the chamber, lowered the stage and re-approached. This accident never happened again.

In my case the same happen already 2 times and the "3View record aborted:Timeout moving knife position to NEAR" situation happen also twice (one of the during the night when nobody was  at work), so I do not feel safe to leave scans overnight ;-(    What if it finally unblock and start cutting again?



3.

You may try to check the contact that Joergen suggested (I am don't know where it is myself). I would just keep monitoring the system and if the same accidents occur 3 times or more start to worry. If the event is not reproducible it will be very difficult to tackle the problem and you might ending up in replacing part after part.

Is anybody from GATAN on this mailing list? How much does it cost to invite GATAN specialist to diagnose the problem? As far as I know we do not have GATAN representative in Poland and the Zeiss who sold us the setup wash their hands of anything connected with GATAN devices  ;-(



Thanks

Gosia

Are you on DM2 or GMS3?



Best regards,

Ilya



--------------------------------------------------------------

Ilya Belevich, PhD

Senior researcher, Electron Microscopy Unit

Institute of Biotechnology, PL56 (Viikinkaari 9)

University of Helsinki, 00014 FINLAND

phone: +358-2-941-59497   email: ilya.belevich(AT)helsinki.fi<https://protect-au.mimecast.com/s/DdneCP7yOZtGQnoEf0seNP?domain=helsinki.fi>

web:   http://www.biocenter.helsinki.fi/~ibelev<https://protect-au.mimecast.com/s/L0mSCQnzP0tA7L3jSMbD63?domain=biocenter.helsinki.fi>



From: 3view-bounces at open-em.org<mailto:3view-bounces at open-em.org> [mailto:3view-bounces at open-em.org] On Behalf Of Malgorzata Sliwinska
Sent: Wednesday, August 22, 2018 6:20 PM
To: Mailing list for Gatan 3View Serial Block-Facing <3view at open-em.org<mailto:3view at open-em.org>>
Subject: Re: [3view] knife moving problem



Hi Joergen,

Thank you for your suggestions. Would it be possible for you to send me a picture showing what you mean by electric wire which bend over time? I would like to try to bend it back. I hope it is not a motor as I am able to cut, after error I start new scan and it is working fine for two hours now... How to check if it is because of cutter piezo?

Generally can I do everything myself  or I need to call Gatan?

Thanks a lot

Gosia















Małgorzata Śliwińska, PhD

Laboratory of Imaging Tissue Structure and Function

Nencki Institute of Experimental Biology Polish Academy of Sciences

3 Pasteur Street, 02-093 Warsaw, Poland

Phone: +48 22 589 2502

www.nencki.gov.pl<https://protect-au.mimecast.com/s/jJsYCROAQotmNXRqSOfBMI?domain=nencki.gov.pl>

W dniu 2018-08-22 16:34, Joergen Kornfeld napisał(a):

Hi Gosia,

it is a bit hard to diagnose for me from the description, but it might be the "cutter motor" or the detection mechanism that tests in which position the cutter arm is. For imaging, the whole knife holding arm is moved away from the sample (this may be what you mean by CLEAR position), but then returned over the sample for cutting. This position is detected by an electric wire contact, which might bend over time so that it is not functional anymore and the cutter arm can get stuck (just bending it back would fix it then). Another possibility is that the cutter motor itself is broken (the entire arm with the knife does not move anymore, needs replacement of the motor) or that the knife piezo does not work anymore (the diamond cannot slide over the sample anymore during approach, needs replacement of the cutter piezo).

Maybe this helps,

Joergen



Am 8/22/2018 um 4:14 PM schrieb Małgorzata Śliwińska:

Dear All,

Recently I have experienced three times a problem with knife moving in 3View.

I the middle of the scan the knife stopped cutting. The pictures were collected, I hear clicking from the knife driver, so it seems that everything is working but 20 consecutive pictures were the same, no cutting. My impression is that the table with the sample was moving up but the knife was blocked in CLEAR position. When it happened I opened the chamber, lower the stage and approached the sample again (because I was afraid that if the knife unblock finally, it will hit the sample and be damaged). After restart the scan continued properly.

Today it happen again, but this time the scan was stopped automatically and I received the message: "3View record aborted:Timeout moving knife position to NEAR"

Also from one year on I have been observing, that while cutting with open chamber, around the 12-17th slice I hear the sound like  "boing" of the metal spring (I did not observed any blockade with the open chamber but I suspect that it can be the problem of the spring responsible for moving the knife arm...)

Does anybody experience similar problem? Does it concern a part easily exchangeable or is it a bigger problem? I am afraid to leave 3View working unattended till I fix this problem ;-(

I would be grateful for any advice

Best regards

Gosia









Małgorzata Śliwińska, PhD

Laboratory of Imaging Tissue Structure and Function

Nencki Institute of Experimental Biology Polish Academy of Sciences

3 Pasteur Street, 02-093 Warsaw, Poland

Phone: +48 22 589 2502

www.nencki.gov.pl<https://protect-au.mimecast.com/s/jJsYCROAQotmNXRqSOfBMI?domain=nencki.gov.pl>



_______________________________________________

3view mailing list

3view at open-em.org<mailto:3view at open-em.org>

http://open-em.org/mailman/listinfo/3view<https://protect-au.mimecast.com/s/CVVyCVAGXPtwM85oHQIFsv?domain=open-em.org>



_______________________________________________

3view mailing list

3view at open-em.org<mailto:3view at open-em.org>

http://open-em.org/mailman/listinfo/3view<https://protect-au.mimecast.com/s/CVVyCVAGXPtwM85oHQIFsv?domain=open-em.org>


_______________________________________________
3view mailing list
3view at open-em.org<mailto:3view at open-em.org>
http://open-em.org/mailman/listinfo/3view<https://protect-au.mimecast.com/s/CVVyCVAGXPtwM85oHQIFsv?domain=open-em.org>


<stage error.JPG>
<microscope comunicat.JPG>
<server comunicate.JPG>
<3View comunicat.JPG>