[3view] knife moving problem

Louise Hughes lhughes at brookes.ac.uk
Thu Aug 23 09:08:07 CEST 2018


Dear All,

I am on GMS3 and have had the same issues with the timeout message several
times and also no slicing upon approach for several microns. It is a
periodic error and was usually corrected by restarting the system. I think
it is a communication issue. It is worth talking to Gatan about this. After
we had several incidents they sent an engineer out and they replaced one of
the USB ports or connectors (if I recall correctly). It doesn't seem to
have happened since, although some of the other communication issues with
the Zeiss computer still seem to occur (this may not be a Gatan issue,
however).

This seems to be a separate issue than the noises you are hearing and the
seizure of the system. Have you contacted Gatan?

I hope this is helpful.

Regards,
Louise

On 23 August 2018 at 06:29, Belevich, Ilya <ilya.belevich at helsinki.fi>
wrote:

> Hi Gosia,
>
> I agree with Joergen, it is difficult to target the problem. Especially,
> since it happened only few times during the year. Your events also may be
> due to certain accidents, such as miscommunication between DM and the knife
> controller (perhaps, in the second case with the error message: 3View
> record aborted:Timeout moving knife position to NEAR).
>
> With the other symptoms, I also had a weird case when I suppose to cut but
> nothing was going on for several microns. Basically, I just opened the
> chamber, lowered the stage and re-approached. This accident never happened
> again.
>
> You may try to check the contact that Joergen suggested (I am don’t know
> where it is myself). I would just keep monitoring the system and if the
> same accidents occur 3 times or more start to worry. If the event is not
> reproducible it will be very difficult to tackle the problem and you might
> ending up in replacing part after part.
>
>
>
> Are you on DM2 or GMS3?
>
>
>
> Best regards,
>
> Ilya
>
>
>
> --------------------------------------------------------------
>
> Ilya Belevich, PhD
>
> Senior researcher, Electron Microscopy Unit
>
> Institute of Biotechnology, PL56 (Viikinkaari 9)
>
> University of Helsinki, 00014 FINLAND
>
> phone: +358-2-941-59497   email: ilya.belevich(AT)helsinki.fi
>
> web:   http://www.biocenter.helsinki.fi/~ibelev
>
>
>
> *From:* 3view-bounces at open-em.org [mailto:3view-bounces at open-em.org] *On
> Behalf Of *Malgorzata Sliwinska
> *Sent:* Wednesday, August 22, 2018 6:20 PM
> *To:* Mailing list for Gatan 3View Serial Block-Facing <3view at open-em.org>
> *Subject:* Re: [3view] knife moving problem
>
>
>
> Hi Joergen,
>
> Thank you for your suggestions. Would it be possible for you to send me a
> picture showing what you mean by electric wire which bend over time? I
> would like to try to bend it back. I hope it is not a motor as I am able to
> cut, after error I start new scan and it is working fine for two hours
> now... How to check if it is because of cutter piezo?
>
> Generally can I do everything myself  or I need to call Gatan?
>
> Thanks a lot
>
> Gosia
>
>
>
>
>
>
>
>
>
>
>
>
>
>
>
> Małgorzata Śliwińska, PhD
>
> Laboratory of Imaging Tissue Structure and Function
>
> Nencki Institute of Experimental Biology Polish Academy of Sciences
>
> 3 Pasteur Street, 02-093 Warsaw, Poland
>
> Phone: +48 22 589 2502
>
> www.nencki.gov.pl
>
> W dniu 2018-08-22 16:34, Joergen Kornfeld napisał(a):
>
> Hi Gosia,
>
> it is a bit hard to diagnose for me from the description, but it might be
> the "cutter motor" or the detection mechanism that tests in which position
> the cutter arm is. For imaging, the whole knife holding arm is moved away
> from the sample (this may be what you mean by CLEAR position), but then
> returned over the sample for cutting. This position is detected by an
> electric wire contact, which might bend over time so that it is not
> functional anymore and the cutter arm can get stuck (just bending it back
> would fix it then). Another possibility is that the cutter motor itself is
> broken (the entire arm with the knife does not move anymore, needs
> replacement of the motor) or that the knife piezo does not work anymore
> (the diamond cannot slide over the sample anymore during approach, needs
> replacement of the cutter piezo).
>
> Maybe this helps,
>
> Joergen
>
>
>
> Am 8/22/2018 um 4:14 PM schrieb Małgorzata Śliwińska:
>
> Dear All,
>
> Recently I have experienced three times a problem with knife moving in
> 3View.
>
> I the middle of the scan the knife stopped cutting. The pictures were
> collected, I hear clicking from the knife driver, so it seems that
> everything is working but 20 consecutive pictures were the same, no
> cutting. My impression is that the table with the sample was moving up but
> the knife was blocked in CLEAR position. When it happened I opened the
> chamber, lower the stage and approached the sample again (because I was
> afraid that if the knife unblock finally, it will hit the sample and be
> damaged). After restart the scan continued properly.
>
> Today it happen again, but this time the scan was stopped automatically
> and I received the message: "3View record aborted:Timeout moving knife
> position to NEAR"
>
> Also from one year on I have been observing, that while cutting with open
> chamber, around the 12-17th slice I hear the sound like  "boing" of the
> metal spring (I did not observed any blockade with the open chamber but I
> suspect that it can be the problem of the spring responsible for moving the
> knife arm...)
>
> Does anybody experience similar problem? Does it concern a part easily
> exchangeable or is it a bigger problem? I am afraid to leave 3View working
> unattended till I fix this problem ;-(
>
> I would be grateful for any advice
>
> Best regards
>
> Gosia
>
>
>
>
>
>
>
>
>
> Małgorzata Śliwińska, PhD
>
> Laboratory of Imaging Tissue Structure and Function
>
> Nencki Institute of Experimental Biology Polish Academy of Sciences
>
> 3 Pasteur Street, 02-093 Warsaw, Poland
>
> Phone: +48 22 589 2502
>
> www.nencki.gov.pl
>
>
>
> _______________________________________________
>
> 3view mailing list
>
> 3view at open-em.org
>
> http://open-em.org/mailman/listinfo/3view
>
>
>
> _______________________________________________
>
> 3view mailing list
>
> 3view at open-em.org
>
> http://open-em.org/mailman/listinfo/3view
>
>
> _______________________________________________
> 3view mailing list
> 3view at open-em.org
> http://open-em.org/mailman/listinfo/3view
>
>


-- 
Dr. Louise Hughes
Bio-Imaging Unit Researcher and  Microscopist
Department of Biological and Medical Sciences,
Oxford Brookes University,
Oxford, OX3 0BP, UK
01865 484351

Brookes e-mail: lhughes at brookes.ac.uk
Personal website: http://www.miniaturehorizons.com
Twitter: https://twitter.com/DrLouiseCHughes