Joergen Kornfeld Joergen.Kornfeld at mpimf-heidelberg.mpg.de
Wed Aug 22 16:34:33 CEST 2018


Hi Gosia,

it is a bit hard to diagnose for me from the description, but it might 
be the "cutter motor" or the detection mechanism that tests in which 
position the cutter arm is. For imaging, the whole knife holding arm is 
moved away from the sample (this may be what you mean by CLEAR 
position), but then returned over the sample for cutting. This position 
is detected by an electric wire contact, which might bend over time so 
that it is not functional anymore and the cutter arm can get stuck (just 
bending it back would fix it then). Another possibility is that the 
cutter motor itself is broken (the entire arm with the knife does not 
move anymore, needs replacement of the motor) or that the knife piezo 
does not work anymore (the diamond cannot slide over the sample anymore 
during approach, needs replacement of the cutter piezo).

Maybe this helps,

Joergen


Am 8/22/2018 um 4:14 PM schrieb Małgorzata Śliwińska:
>
> Dear All,
>
> Recently I have experienced three times a problem with knife moving in 
> 3View.
>
> I the middle of the scan the knife stopped cutting. The pictures were 
> collected, I hear clicking from the knife driver, so it seems that 
> everything is working but 20 consecutive pictures were the same, no 
> cutting. My impression is that the table with the sample was moving up 
> but the knife was blocked in CLEAR position. When it happened I opened 
> the chamber, lower the stage and approached the sample again (because 
> I was afraid that if the knife unblock finally, it will hit the sample 
> and be damaged). After restart the scan continued properly.
>
> Today it happen again, but this time the scan was stopped 
> automatically and I received the message: "3View record 
> aborted:Timeout moving knife position to NEAR"
>
> Also from one year on I have been observing, that while cutting with 
> open chamber, around the 12-17th slice I hear the sound like  "boing" 
> of the metal spring (I did not observed any blockade with the open 
> chamber but I suspect that it can be the problem of the spring 
> responsible for moving the knife arm...)
>
> Does anybody experience similar problem? Does it concern a part easily 
> exchangeable or is it a bigger problem? I am afraid to leave 3View 
> working unattended till I fix this problem ;-(
>
> I would be grateful for any advice
>
> Best regards
>
> Gosia
>
>
>
>
> Małgorzata Śliwińska, PhD
> Laboratory of Imaging Tissue Structure and Function
> Nencki Institute of Experimental Biology Polish Academy of Sciences
> 3 Pasteur Street, 02-093 Warsaw, Poland
> Phone: +48 22 589 2502
> www.nencki.gov.pl
>
>
> _______________________________________________
> 3view mailing list
> 3view at open-em.org
> http://open-em.org/mailman/listinfo/3view